Ge campaign post calibration analysis

The efficiency curves shown below apply to NSCL Expts. 02004, 01027, and 01015, which were completed during the period April 14 thru May 10, 2003. Twelve 32-segment Ge detectors from the MSU SeGA array were placed in two rings around the DSSD implantation detector.

Germanium array geometry

The germanium detector geometries as entered into the Monte-Carlo N-Particle (MCNP) transport program are shown given in the links below. The positive X direction is downstream of the DSSD implantation detector, the positive Y direction corresponds to the South direction in the N3 vault.

XZ plane, Y offset +1.5 cm YZ plane, X offset -2.8 cm YZ plane, X offset +3.6 cm

Photos of the Ge detector arrangement can be found at the following link:

groups.nscl.msu.edu/beta/BetaDecay/calorimeter.html

Individual detector efficiencies

Source: SRM-4275-C69 for absolute calibration

Source: Co-56 prepared at FSU for relative calibration to 3.2 MeV

Data were collected using both an MCA and the NSCL Spectrodaq system. For the Spectrodaq measurements, independent measurements were made for each detector.

Ge01 #73444 Ge02 #73486 Ge03 #73177
Ge04 #73452 Ge05 #73446 Ge06 #73494
Ge09 #73288 Ge10 #73471 Ge11 #73445
Ge12 #73373 Ge13 #73474 Ge14 #73375


Array peak efficiency

The peak efficiency for the total 12-detector array is given below. The data were fitted to an MCNP simulation for a point-source geometry, and the overall agreement between 150 keV and 3.5 MeV is excellent. The posted equation is the result of a trendline fit to the MCNP simulation points (log Efficiency vs. log Energy) using a fifth-order polynomial function.

Ge array peak efficiency curve


MCNP Simulation File

The MCNP source code to calculate the efficiency curve for the point source geometry can be downloaded from the link below. To save the file on your local disk, click on the link with the right mouse button and select "Save Target As..."

MCNP file 01015P.i


updated 30may03 by pfm