High Resolution Array

Detector Database

Detector Homepage
All Detector Comments:
Part # Date Author Comments
2006-01-24 D.Henzlova one bad preamp. found - preamp. box replaced with one from telescope 15
2006-11-11 Vlad Henzl dE 2297-7 mounted into the telescope 17
2007-02-02 F. Last 128 Character Max.
2007-08-14 F. Last 128 Character Max.
103 2007-07-30 A. Sanetullaev calibrated at WMU
2085-11 2006-06-02 D.Henzlova taken by Bob Charity for Texas experiment
2085-12 2005-05-31 F. Delaunay Could not be biased. Shipped back to Micron.
2085-12 2006-02-09 W. Lynch Rejected, no longer owned by HiRA Collaboration
2085-14 2006-09-30 Vlad Henzl ef32 discharging
2085-14 2006-11-02 Vlad henzl Back from Micron => detector cleaned and EF32 bonded together with EF31, bond pad of EF32 used for wirebond to guardring, which can be now biased.
2085-14 2006-11-11 Vlad Henzl Many wirebonds at back side broken : EB9-1, 11-3, 12-1, 13-2, 14-2, 15-3, 16-1, 17-1, 18-2, 19-3, 20-2, 21-2, 23-2, 24-1, 25-1, 26-1, 27-1, 28-2
2085-2 2006-06-02 D.Henzlova taken by Bob Charity for Texas experiment
2085-2 2006-11-07 D.Henzlova strip Eb17 had all 3 wirebonds unattached - rebonded by John Yurkon
2085-6 2006-06-24 Vlad Henzl detector involved in a incident when chamber vented while detectors biased. So far seems OK.
2085-6 2007-05-15 C. Hollek Sent to Washington University [wire bonds were in good condition]
2085-6 2007-08-07 C. Hollek Received form Washington University, wire bonds in good condition
2085-7b 2006-10-31 D.Henzlova back from Micron => detector cleaned and EF32 bonded together with EF31, bond pad of EF32 used for wirebond to guardring, which can now be biased
2086-7 2006-06-02 D.Henzlova taken by Bob Charity for Texas experiment
2113-5 2006-03-09 D.Henzlova golden film on the back side of the detector seems to be slightly unattached from the detector surface
2113-6 2007-03-27 D.Henzlova used in 03045 experiment in telescope 13 in replacement of 2403-4, which was at Micron at that time
2113-6 2006-11-18 V.Henzl during testing inside 4pi at 270V very noisy - extremely underbiased. Increasing biaas to 410V clears up the situation
2113-7 2006-01-06 D. Henzlova Bob Charity reports 7% lower energy of back middle strips (10-20) calibration than in the corresponding front strips - check if it is the problem of detector or electronics
2113-7 2006-07-28 Vlad Henzl Strong discharge on EF1+32 and EB1+32 already at 300 V => sent to Micron for repair
2113-7 2006-11-02 Vlad Henzl Back from Micron => detector cleaned and EF32 bonded together with EF31, bond pad of EF32 used for wirebond to guardring, which can be now biased.
2113-7 2006-11-02 Vlad Henzl When guardring not biased, strong discharge over all strips, when GR biased, detector stable and working well up to 425 Volts
2260-2 2005-05-25 S. Tornga 2 broken wire bonds
2260-2 2006-11-11 Vlad Henzl Mounted into Tel. 16
2260-5 2007-10-06 D. Henzlova detector removed from tel.7 and mounted in telescope 15
2266-4 2007-05-15 C. Hollek Sent to Washington University [wire bonds were in good condition]
2266-4 2007-08-07 C. Hollek Received form Washington University, wire bonds in good condition
2266-6 2006-06-08 D.Henzlova sent to WU for Texas experiment
2266-9 2006-06-02 D.Henzlova taken by Bob Charity for Texas experiment
2297-10 2006-06-08 D.Henzlova prepared to be sent to WU for Texas experiment
2297-10 2007-05-15 C. Hollek Sent to Washington University [wire bonds were in good condition]
2297-10 2007-08-07 C. Hollek Received form Washington University, wire bonds in good condition
2297-3 2007-05-15 C. Hollek Sent to Washington University [wire bonds were in good condition]
2297-3 2007-08-07 C. Hollek Received form Washington University, wire bonds in good condition
2297-9 2007-05-15 C. Hollek Sent to Washington University [wire bonds were in good condition]
2297-9 2007-08-07 C. Hollek Received form Washington University, wire bonds in good condition
2344-03 2007-05-15 Colby Hollek Sent to Washington University [forgot to check wire bonds]
2344-03 2007-08-07 C. Hollek Received form Washington University, wire bonds in good condition
2344-5 2006-10-31 D.Henzlova back from Micron => detector cleaned and EF32 bonded togehter with EF31, bond pad of EF32 used for wirebond to guardring, which can now be biased
2393-14 2005-03-25 F. Delaunay Detector tilted in frame. Corner pin between the cables can not be fully inserted.
2393-14 2006-02-09 D.Henzlova after experiment 02023 it was observed that wire bonds on the back side of the detector popped up. Shipped back to Micron
2393-14 2006-03-01 D.Henzlova arrived back from Micron after repair - wire bonds were rebonded; NEEDS TO BE TESTED
2393-14 0000-00-00 D.Henzlova Arrived back from WU after their TEXAS experiment
2393-14 2007-05-15 C. Hollek Sent to Washington University [wire bonds were in good condition]
2393-14 2007-08-07 C. Hollek Received form Washington University, wire bonds in good condition
2393-14b 2005-06-17 F. Delaunay This detector has number 2399-14 in IU database.
2393-14b 2006-06-08 D.Henzlova prepared to be sent to WU for Texas experiment
2393-14b 2006-09-22 D.Henzlova due to numerous unattached wire bonds sent to Micron for rebonding
2393-14b 2006-10-23 Vlad Henzl received from Micron - all bonds now look OK
2393-14b 2006-10-23 Vlad Henzl EF 27 heavily discharging, gentle troubles at EF1+32 observable too
2403-2 2005-06-08 F. Delaunay Front evens & odds are inverted: strip1 -> pin1, strip32 -> pin34, pins 2 & 33=gnds.
2403-2 2005-06-08 F. Delaunay Needs to be tested.
2403-2 2006-09-30 Vlad + Jenny strong discharge on EF3, 6, 7 ... sent to Micron for repair
2403-2 2006-11-02 Vlad Henzl Back from Micron => cleaned + EF31 and EF32 bonded together, guard ring bonded instead of EF32
2403-2 2006-11-03 Vlad Henzl + Jenny Lee EF6 strongly discharging already at app. 15 Volts
2403-4 2005-06-13 F. Delaunay. Needs to be tested.
2403-4 2006-01-10 D.Henzlova too high leakage current (~ 9uA), shipped back to Micron
2403-7 2006-06-08 D.Henzlova prepared to be sent to WU for Texas experiment
2403-7 2007-05-15 C. Hollek Sent to Washington University [wire bonds were in good condition]
2403-7 2007-08-07 C. Hollek Received form Washington University, wire bonds in good condition
2573-3 2007-08-07 D.Henzlova new assembly, recieved on 08/03/2007 from Micron - needs to be tested
280SAA 2007-07-30 A. Sanetullaev calibrated at WMU
281SAA 2006-03-11 Alisher Sanetullaev, Jenny Lee Corner peeling off
281SAA 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:5 CsI-0)
281SAA 2007-07-30 D.Henzlova calibrated at WMU
282 2007-07-30 A. Sanetullaev calibrated at WMU
285SAA 2006-03-11 128 Character Max.
285SAA 2006-03-11 Alisher Sanetullaev, Jenny Lee hole at adge
285SAA 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:6 CsI-0)
286SAA 2006-03-11 Alisher Sanetullaev, Jenny Lee Corner peeling off
286SAA 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:6 CsI-1)
287SAA 2007-07-30 A. Sanetullaev calibrated at WMU
288SAA 2006-03-11 Alisher Sanetullaev, Jenny Lee hole at edge
288SAA 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:5 CsI-1)
289SAA 2007-07-30 A. Sanetullaev calibrated at WMU
290 2007-07-30 A. Sanetullaev calibrated at WMU
800SSW 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:6 CsI-2)
837 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:9 CsI-0)
837 2007-07-30 A. Sanetullaev calibrated at WMU
838 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:4 CsI-0)
838 2007-05-24 C. Hollek Removed and checked for damage from telescope 9, found to be in good condition.
839 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:3 CsI-2)
839 2007-07-30 D.Henzlova calibrated at WMU
840 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:3 CsI-3)
841 2007-04-09 Alisher Sanetullaev Needs to be decoupled
841 2007-06-15 Alisher Sanetullaev Not able to decouple
842 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:1 CsI-0)
843 2007-04-09 Alisher Sanetullaev Decoupled for for polishing and reglueing
843 2007-06-14 Alisher Sanetullaev Tested with cosmic rays, to be tested in WMU with protons.
844 2006-03-09 Jenny Lee take out for repair
844 2006-03-09 Jenny Lee small holes
844 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:11 CsI-1)
846 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:2 CsI-2)
846 2007-05-24 C. Hollek Removed and checked for damage from telescope 7, found to be in good condition.
847SSW 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:8 CsI-2)
847SSW 2007-05-24 C. Hollek Removed and checked for damage from telescope 7, found to be in good condition.
848 2007-07-30 Sanetullaev calibrated at WMU
849 2007-04-09 Alisher Sanetullaev Decoupled for for polishing and reglueing
849 2007-06-15 Alisher Sanetullaev Tested with cosmic rays, to be tested in WMU with protons.
850 2007-04-09 Alisher Sanetullaev Decoupled for for polishing and reglueing
850 2007-06-15 Alisher Sanetullaev Tested with cosmic rays, to be tested in WMU with protons.
850 2007-07-30 A. Sanetullaev calibrated at WMU
851 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:16 CsI-3)
853 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:10 CsI-0)
854 2006-03-09 Jenny Lee take out for repair (small holes)
854 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:11 CsI-3)
854 2007-07-30 A. Sanetullaev calibrated at WMU
855 2007-06-15 C. Hollek Decoupled , glued and tested before Mike’s experiment - November 06
856SSW 2007-06-15 Alisher Sanetullaev Slightly damaged while decoupling, rough spot grinded, ok now.
856SSW 2007-06-15 Alisher Sanetullaev Tested with cosmic rays, to be tested in WMU with protons.
857 2007-06-15 C. Hollek Decoupled , glued and tested before Mike’s experiment - November 06
858 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:4 CsI-1)
858 2007-05-24 C. Hollek Removed and checked for damage from telescope 9, found to be in good condition.
859 2007-07-30 Sanetullaev calibrated at WMU
860SSW 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:8 CsI-1)
860SSW 2007-07-30 Sanetullaev calibrated at WMU
861 2007-06-15 Alisher Sanetullaev Tested with cosmic rays, to be tested in WMU with protons.
862 2007-06-15 Alisher Sanetullaev Tested with cosmic rays, to be tested in WMU with protons.
863 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:19 CsI-2)
863 2007-07-30 A. Sanetullaev calibrated at WMU
865 2007-06-15 C. Hollek Decoupled , glued and tested before Mike’s experiment - November 06
865 2007-07-30 A. Sanetullaev calibrated at WMU
866 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:2 CsI-1)
867 2006-03-11 Alisher Sanetullaev, Jenny Lee pealing off at corner
867 2007-04-09 Alisher Sanetullaev Decoupled for for polishing and reglueing
868 2007-04-09 Alisher Sanetullaev Decoupled for for polishing and reglueing
868 2007-06-15 Alisher Sanetullaev Tested with cosmic rays, to be tested in WMU with protons.
868 2007-07-30 A. Sanetullaev calibrated at WMU
869 2007-04-09 Alisher Sanetullaev Decoupled for for polishing and reglueing
869 2007-06-15 Alisher Sanetullaev Tested with cosmic rays, to be tested in WMU with protons.
870 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:19 CsI-0)
870 2007-07-30 A. Sanetullaev calibrated at WMU
873 2007-06-15 Alisher Sanetullaev Tested with cosmic rays, to be tested in WMU with protons.
873 2007-07-30 A. Sanetullaev calibrated at WMU
874SSW 2006-03-11 Alisher Sanetullaev, Jenny Lee corner of the mylar foil slightly breached.
874SSW 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:5 CsI-3)
874SSW 2007-07-30 Sanetullaev calibrated at WMU
875 2007-04-09 Alisher Sanetullaev Decoupled for for polishing and reglueing
875 2007-06-15 Alisher Sanetullaev Tested with cosmic rays, to be tested in WMU with protons.
875 2007-07-30 A. Sanetullaev calibrated at WMU
877 2007-04-09 Alisher Sanetullaev Decoupled for for polishing and reglueing
877 2007-06-15 Alisher Sanetullaev Tested with cosmic rays, to be tested in WMU with protons.
877 2007-07-30 A. Sanetullaev calibrated at WMU
878 2006-03-11 Alisher Sanetullaev, Jenny Lee corner of the mylar foil slightly breached.
878 2007-07-30 A. Sanetullaev calibrated at WMU
879 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:19 CsI-3)
879 2007-07-30 A. Sanetullaev calibrated at WMU
880 2007-04-09 Alisher Sanetullaev Decoupled for for polishing and reglueing
880 2007-06-15 Alisher Sanetullaev Tested with cosmic rays, to be tested in WMU with protons.
881 2007-07-30 A. Sanetullaev calibrated at WMU
882 2007-07-30 A. Sanetullaev calibrated at WMU
883 2007-04-09 Alisher Sanetullaev Decoupled for for polishing and reglueing
883 2007-06-15 Alisher Sanetullaev Tested with cosmic rays, to be tested in WMU with protons.
884 2006-03-09 Jenny Lee take out for repair (small holes)
884 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:11 CsI-0)
884 2007-07-30 A. Sanetullaev calibrated at WMU
885 2007-06-15 C. Hollek Decoupled , glued and tested before Mike’s experiment - November 06
888 2006-02-02 D.Henzlova cabling of CsI inside the telescope is rotated, i.e. CsI-0, 1, 2, 3 go to channels 1, 2, 3, 0, respectivley - needs to be considered in the SpecTcl mapping
888 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:2 CsI-0)
889 2007-04-09 Alisher Sanetullaev Decoupled for for polishing and reglueing
889 2007-06-15 Alisher Sanetullaev Tested with cosmic rays, to be tested in WMU with protons.
889 2007-06-15 C. Hollek Decoupled , glued and tested before Mike’s experiment - November 06
890 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:0 CsI-2)
891 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:13 CsI-1)
891 2007-07-30 A. Sanetullaev calibrated at WMU
892 2006-01-24 D.Henzlova one bad preamp. found - preamp box replaced with the box from telescope 15
892 2007-06-15 Alisher Sanetullaev Tested with cosmic rays, to be tested in WMU with protons.
892 2007-07-30 A. Sanetullaev calibrated at WMU
893 2007-05-24 C. Hollek Removed and checked for damage from telescope 9, found to be in good condition.
895 2007-07-30 A. Sanetullaev calibrated at WMU
897SSW 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:5 CsI-2)
897SSW 2007-07-30 A. Sanetullaev calibrated at WMU
899 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:14 CsI-1)
899 2007-05-24 C. Hollek Removed and checked for damage from telescope 7, found to be in good condition.
900 2006-03-11 Alisher Sanetullaev, Jenny Lee corner of the mylar foil slightly breached
900 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:14 CsI-3)
900 2007-07-30 A. Sanetullaev calibrated at WMU
901 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:1 CsI-3)
902 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:13 CsI-2)
903 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:10 CsI-3)
903 2007-05-24 C. Hollek Removed and checked for damage from telescope 7, found to be in good condition.
905 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:3 CsI-0)
907 2006-01-24 D.Henzlova photodiode cannot be biased, preamp. O.K. - check connection to preamp., eventually exchange photodiode
907 2006-03-01 D.Henzlova, A.Sanetullaev all CsI crystals can be biased and "see" cosmics - probably the cable connection between preamp. and photodiode was loose before
907 2006-03-11 Alisher Sanetullaev, Jenny Lee Holes in the mylar cover
907 2007-05-17 C. Hollek` Configuration prior to November 06 (Telescope:3 CsI-1)
907 2007-07-30 A. Sanetullaev calibrated at WMU
908 2007-07-30 D.Henzlova calibrated at WMU
910 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:16 CsI-2)
911 2007-06-15 C. Hollek Decoupled , glued and tested before Mike’s experiment - November 06
912 2006-03-09 Jenny Lee take out for repair
912 2006-03-09 Jenny Lee samll holes
912 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:11 CsI-2)
913 2007-06-14 Alisher Sanetullaev Tested with cosmic rays, to be tested in WMU with protons.
914 2006-03-11 Alisher Sanetullaev, Jenny Lee edge slight, y breached
914 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:9 CsI-2)
915 2007-07-30 A. Sanetullaev calibrated at WMU
918 2006-01-24 D.Henzlova photodiode cannot be biased, preamp. O.K. - check connection to preamp., event. exchange
918 2006-03-02 D.Henzlova, A.Sanetullaev all CsI crystals in tele. 17 are working - probably the cable connection between preamp. and photodiode was loose before
918 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:17 CsI-1)
919 2006-03-11 Alisher Sanetullaev, Jenny Lee edge slightly breached
919 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:14 CsI-0)
919 2007-07-30 A. Sanetullaev calibrated at WMU
920 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:9 CsI-1)
920 2007-05-24 C. Hollek Removed and checked for damage from telescope 9, found to be in good condition.
921 2006-03-11 Alisher Sanetullaev, Jenny Lee corner of the mylar foil slightly breached.
921 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:14 CsI-2)
921 2007-07-30 D.Henzlova calibrated at WMU
923 2007-04-09 Alisher Sanetullaev Decoupled for for polishing and reglueing
923 2007-06-15 Alisher Sanetullaev Tested with cosmic rays, to be tested in WMU with protons.
924 2007-06-15 C. Hollek Decoupled , glued and tested before Mike’s experiment - November 06
924 2007-07-30 A. Sanetullaev calibrated at WMU
925 2007-05-17 C. Hollek Configuration prior to November 06 (Telescope:0 CsI-1)
925 2007-07-30 A. Sanetullaev calibrated at WMU
zSMWtest 2006-12-18 F. Last test
zSMWtest 2006-12-18 F. Last test